The situation described (of a shift in the mean fluorescence intensity of
the whole population of cells on aantibody staining) is quite common.

In my view, it would be incorrect (indeed, meaningless) to subtract away
the isotype control fluorescence and then to describe the remaining cells
as 'positive'. The whole population is positive, the observation that the
positive and negative populations overlap is irrelevant. 

In this type of experiment, you are not trying to identify positive and
negative sub-sets but looking at the expression of an antigen across the
whole population.

In these experiments, I always express the results as either the mean or
the median fluorescence intensity of the whole population. 

It could be legitimate to subtract the MFI of the isotype control to
express the result as the increase in MFI due to antibody binding. 