A major problem probably is the actual number of binding sites for your
staining reagent on each phage particle. This might be below the detection
limit of your machine. A second problem might be the staining reagent. If
it is not an antibody, you may want to check if the affinity is
sufficiently high (or the koff sufficiently low) to prevent detachment
during the test.

Saverio Alberti
Head, Unit of Experimental Oncology
Department of Cell Biology and Oncology
Consorzio Mario Negri Sud
66030 Santa Maria Imbaro (Chieti), Italy
Phone: (39) 872-570.293
FAX: (39) 872-578.240
E-mail: alberti@cmns.mnegri.it


On Wed, 3 Feb 1999, Luca Battistini & Giovanna Borsellino wrote:

> 
> 
> Dear cytometrists,
> 
> 	We are trying to sort filamentous phage on a Cytomation MoFlo. Phages are
> pretty tiny, being around 1 micron in length and less than 0.07 micron in
> width. We can actually see it and sort it, but so far have been unable to
> stain it.
> 
> Any ideas will be greatly appreciated!
> 
> Cheers
> 
> Giovanna Borsellino
> IRCCS S.Lucia
> Rome, Italy
> 
> 