With regard to Gene's ??, the following.

Fig 1 in the paper clearly shows that the bimane and Ho42 spectra overlap
alomst completely, so your plan to use a 440LP DC and a 450/50BP will fail
completely. This was the reason we had to go the phase-resolved route.

>From reading the literature we found several probes (mentioned in the
article) that have been used for GSH/thiol analyses. However Hedley found
that bimane was the most specific for GSH. If you are not that concerned
about specifity you can probably use CMFDA and accept "some" co-staining of
other -SH containing molecules.

Indeed, providing a solution for the overlapping spectra was the central
issue in the manuscript.

Regards,

Jan




>Colleagues,

>My question concerns the Phase-Resolved flow cytometry paper
>by Jan Keij et al. in the january Cytometry.

>I had thought people were routinely utilizing Hoechst/MCB to examine
>glutathione and relative DNA content and had hoped to shortly do this
>myself on our FACStar Plus utilizing UV excitation and collection
>across a 440LP dichroic to a standard FITC filter fot the MCB and a
>450/50 for the Hoechst.

>Is monobromobimane much more specific for GSH and its overlapping
>emission profile relative to Hoechst the central issue here?

Gene/UCONN Health