Re: Dendritic cells

From: Steve Perfetto (sperfetto@hiv.hjf.org)
Date: Wed Jan 28 1998 - 10:44:48 EST


Dr Lawry,

We have sorted and characterized both DC's from skin and the so called
macrophage "derived" DC's. The most striking characteristic in both of these
cell types is the  surface expression of HLA-DR.  In which the DR expression on
skin DC's is much brighter than the M0 derived DC's. Depending on whether the
DC's are conjugated with T-Cells or not will define the physical
characteristics.  Generally, these cells scatter high 90 and mid-scale using
peak forward scatter,  however, the skin DC's are larger and may require tuning
down the 90 scatter.  With all this in mind I recommend staining with the
following combinations;

DR/14-56-20/CD3 : Background staining for the monocyte derived DC's.  As cells
"transform" they should loose the CD14 marker and none should be positive for 20
or 56.  If this is done correctly all of these could be PE because you really do
not need these percentages.  This is our combination for sorting.

DR/80/CD3 : On DC's (unconjugated) they will be positive for DR and 80.

DR/83/CD3 : On DC's (unconjugated) they will be positive for DR and 83.

Stephen P. Perfetto, MS.,MT. (ASCP)
Walter Reed Army Institute of Research
Department of Molecular Diagnostics and Pathogenesis
1600 East Gude Drive
Rockville, MD. 20850
_______________________________________________________________________________
Subject: Dendritic cells
From:    J.Lawry@sheffield.ac.uk at Internet_Gateway
Date:    1/27/98  9:35 AM


Dear All,

We are trying to characterise dendritic cells in stimulated cultures of
PBMCs. We have a panel of potential markers of dendritic cells but are
having problems finding their characteristic scatter patterns for gating
purposes. Back gating gives no clear area. Can anyone help? Where do these
cells lie on a normal FWD vs SSC plot of "whole blood" - in the
LGL/monocyte area, or towards the neutrophils?

Many thanks, John.

 Dr John Lawry,
Institute for Cancer Studies, University Medical School, Beech Hill Road,
Sheffield. S10 2RX.
Tel:0114-271-3023 (lab), 0114-271-3774 (office)
Fax:0114-271-3515
Email: J.Lawry@sheffield.ac.uk
Web page: http://www.shef.ac.uk/uni/academic/A-C/csi/Staff/John/JL.html



This archive was generated by hypermail 2b29 : Wed Apr 03 2002 - 11:51:43 EST