Theory and extrapolation from excitation/emission charts are great first
steps, but empirical observations also have their place in applications
development.

Our experience is that Cy5 works well using Rhodamine 6G dye laser
excitation.  Indeed, Cy5 has become our dye of choice for use with Texas
Red, and we now routinely use it for 4, 5 and 6-color applications instead
of APC.

Our experience is that signal:noise with Cy5 is generally as good as with
APC.  Our most direct comparison of this issue was using a single
biotinylated antibody and comparing multiple sources of Cy5-Streptavidin and
APC-Streptavidin.  In this kind of comparison, the very best APC-SA was less
than 2-fold brighter than the very best Cy5-SA.  Several APC-SA's had
signal:noise ratios worse than any of the Cy5-SA's.  Thus, the source/batch
of the conjugated Streptavidin was more important than whether the
fluorochrome was APC or Cy5.  

Signal:noise ratios for things like directly conjugated anti-murine CD8-Cy5
are equivalent, or better, than those we obtain with direct Fitc or PE with
the same clone.  Importantly, it is ridiculously simple to do your own
conjugations with Cy5 using kits.  CALTAG has conjugated several anti-mouse
clones for us with Cy-5.  These work well, we routinely (like everyday) use
them and I believe CALTAG sells at least some of their clones conjugated to
Cy-5 as standard products.

For us, one of the major benefits of using Cy5 instead of APC is that the
back cross-talk (the APC or Cy5 back into the Texas Red channel) is much,
much less with Cy5 than with APC.  In general, we use about 4-fold less back
compensation (FL3 - FL4) with Cy5 than with APC.

We generally run our dye laser at 595-600 nm, about 125mW.  We use a 630/22
filter for Texas Red and a 670/14 for Cy-5.  If you have a BDIS
FACSTAR/VANTAGE type instrument, it is important to make sure that BD has
the correct dichroic splitters and filters in your Texas Red and APC
channels as they have at times been a bit confused in this area.  Our
experience is also that simultaneous optimization of both the APC/Cy5
channels and the tandem (Cychrome, Red670) channel requires very careful
alignment.  

Susan Sharrow
Experimental Immunology Branch
National Cancer Institute
301-496-2776



 