	I am scratching my head after spending the last hour with clients who want
to use 7-AAD to measure apoptosis and also stain with FITC and PE to
measure various monoclonals.  When we run their cells with the 7-AAD stain
only we get nice differentiation of live, apoptotic and dead cells.  When I
then run a sample stained with FITC only on the 7-AAD program the
fluorescence mesured on the 7-AAd PMT (PMT 3, 625) goes way up.  Of course
it does, FITC needs to be compensated out.  When I go to my color comp
controls, nothing I do alters this problem.  I got so frustrated I turned
up the FL-1 and FL-3 color comp components all the way up and nothing
changed.  So what's up here?  The same "phonomenon" occurred when I ran a
PE only stained sample.  The 7-AAD fluorescence changed considerably from a
sample run with 7-AAD and a sample stained with PE only, with no response
from the color comp controls.  I know the color comp controls work because
earlier in their visit we set up the FITC/PE samples and the color comp did
it's job.
	Any suggestions?  Does what I described make any sense, or am I rambling?
It's very possible because I am at a loss right now.
	Any and all suggestions at how to handle this phenomenon would be
appreciated.