On 24 Nov 1997 17:09:38 U, Alan Stall wrote: > It's not necessary to have compensation if you are using FITC, PE or PerCP in > addition to the APC. However if you use a PE-Cy5 tandem you have to have > crossbeam compensation because the red diode laser will also excite the Cy5 > part to the PE-Cy5 tandem giving a false signal in the FL4 parameter. > > Alan Stall > PharMingen This is unfortunately not true. Even if BDIS and PharMingen are very interested to sell their PerCP to dual laser FACSCalibur users, significant compensation is required between PerCP and APC on this instrument. However, the compensation is different for PerCP and APC when compared to PE-Cy5 and APC which can be used as well. PE-Cy5 conjugates in general have the significant advantage of being far brighter than PerCP conjugates. The only problem with PE-Cy5 is that some manufactures have problems in providing reagents with a lot-to-lot stable ratio of the PE and Cy5 emission. Furthermore, reagents from some but not all manufacturers tend to stick non-specifically to monocytes. Gregor Rothe *********************************************************** Institute for Clinical Chemistry and Laboratory Medicine University of Regensburg D-93042 Regensburg, Germany Tel. +49 (941) 944-6204 Fax +49 (941) 944-6202 E-mail: Gregor.Rothe@KLINIK.UNI-REGENSBURG.DE ***********************************************************