From: Jörg Hildmann (hildh000@goofy.zdv.Uni-Mainz.de)
Date: Tue Sep 30 1997 - 18:08:13 EST


Hello everyone,

I was glad that I found latex particles with are coupled with a fluorochrome
mixture which absorb at 488nm and send out 685nm (Molecular Probes,
Transfluorospheres 488/685nm, 1µm).

But my happiness ended when I tried to measure a triple fluorescent
population of cells labelled which FITC, PE and these 488/685nm beads. These
beads are very well labelled that one particle gives about 3 logs of
fluorescence. In FL2 early 2,5 logs are measured and can be compensated by
10% FL3->FL2. But there are nearly 1,5 logs left in the FL1 (i think that
the excision fluorochrome doesn't give all its absorbed energy to the
emission fluorochrome and emits itself) where I get a really big number of
FL1 negative and 1+ FL3 positive populations (and also each combination of
them).

The FACSsort with CellQuest 1.2 allows only to compensate between FL1/FL2 or
FL2/FL3 - not FL3->FL1 or FL1->FL3. Compensation of FL3->FL1 over
FL3->FL2->FL1 destroys completely FL1 by a strong FL2 signal or no
compensation results in multiple FL1-/FL3+, FL1+/FL3- and FL1+/FL3+
populations which make a good analysis very hard and presentation of results
nearly impossible.

I think there are only two possibilities for me! {1} Who knows how I can
overcome these compensation problem by software (Upgrade on CellQuest 3.1
doesn't solve the problem because FL3->FL1 isn't implemented yet) or
hardware (is there any other detection filter combination available to
prevent huge FL3->FL1 influence). {2} Another type of polystyrene particles
available which haven’t these FL3-> FL1 compensation problem (like the BD
calibrate beads 2µm with PerCP label - PerCP is a monomeric protein and has
nearly no FL3-> FL1 activity - but particles have wrong diameter).

Thanks for all help and good luck for your experiments