From: Steve G. Hilliard (steve@cb.uga.edu)
Date: Mon Sep 08 1997 - 09:53:01 EST


Hi folks,
While I'm certainly no expert in this area, I feel compelled to
point out that it is the fluorescence _intensity_ of fluorescein
that varies with pH;  the peak emission wavelength does NOT change.  See
Dr. Haugland's chapter in the latest Methods in Cell Biology manual (part
B), p 658.  I had to rebut this point in response to a reviewer's 
comments recently, which is the only reason I remember it ;-)

I have seen the same spectral difference with an antibody made here on
campus, but the details of its conjugation were not available.  In this 
case I suspect Dr. Prussin is on the right track;  it's a matter of
FITC =//= fluorescein.  

Regards,
Steve

---------------------------------------------------------------
Steve G. Hilliard        steve@habanero.cb.uga.edu        
Cell Analysis Facility, University of Georgia