Hello, everybody

We have noticed on the list several discussions about surface molecules on
adherant cells and the effect of trypsin. Since we had undertaken a few
number of experiments we thought that it would be interesting for a larger
number of Flowers to have access to some of our results.
Our purpose was to check whether trypsin would alter the quantitation of
various adhesion molecules on adherant cells. Here the cells we used were
Fibroblasts MRC5. When used at appropriate concentration for addherant cells
detachment (we use 0.05% trypsin + 0.02% EDTA for up to 15 minutes at 37°C,
maximum), it will not modify the expression of these adhesion molecules as
shown using the QIFI technology (indirect immunofluorescence staining and
quantitative flow cytometry readings).
As an example quantitation of integrins CD49b,d,e,f and CD29 on fibroblasts
was comparable when detached using either Trypsin -EDTA or Versene (see
table 1 below). One may noticed a variation for CD49d although we should
certainly repeat our experiment. In our hands cell recovery and viability
tends to be poorer using Versene treatment.

Table 1: Comparison of quantitation of surface molecules on fibroblasts MRC5
after either Trypsin or versene cell detachment procedures
Results expressed as quantitative number of bound MoAbs per cell

Monoclonal Antibodies	Trypsin	Versene

Isotypic control IgG2a        	  1,000	    5,100
Isotypic control IgG1	                     3400	    1,000

CD9	                                    545,000	510,400
CD49b	                                    254,500	269,500
CD49d	                                      11,400	  22,500
CD49e	                                    197,300	243,000
CD49f	                                      19,250	  21,100
CD146	                                    244,150	271,750
CD51	                                      59,900	  75,600
CD29	                                    795,495	746,750

Positive control	                 724,850	778,400
(MoAbs Cocktail reacting on human cells)

In another experiment we have quantified integrins on human keratinocytes
after trypsin treatment:
The results were as follows:

Monoclonal antibodies                   MAb molecules bound per cell

CD29                                               700,000
CD49b                                             295,000
CD49d                                                7,500
CD49e                                              46,500
CD49f                                             161,000
CD51                                                41,000
CD9                                             1,340,000

In addition we noticed a clear difference between the quantitation of CD146
on Fibroblast (> 200,000 sites per cell) and on keratinocytes (< 2,000 sites
per cell).

In the case someone is looking for more details and specific conditions for
endothelial cells we suggest the following paper:
Reevaluation of Trypsin-EDTA for Endothelial Cell Detachment before Flow
Cytometry Analysis.
Mutin, M, George, F, Lesaule, G, Sampol, J.
Endothelium, 1996, vol 4, pp 289-295.

Michel Canton, PharmD
Philippe Poncelet, PhD

from BioCytex, Marseille, France