Jan, > We have someone who wants to use a monoclonal antibody conjugated to Cy3. > They have been told you can't use a FACScan as Cy3 won't excite at 488nm. > Any suggestions which instruments can be used. You can indeed use Cy3 on a FACScan. We have looked at a dozen or more in-house conjugates. While Cy3 would *prefer* to be excited at around 550nm, or even at 515nm, you can still get enough signal using 488nm if the antibody is suitable. Expect up to one decade less signal than for a comparable PE conjugate. \ / < The Flow Systems Laboratory Frank Battye, \__/ <<<<< The Walter & Eliza Hall Institute Dora Kaminaris,----------!!<<<<<<<< Victoria 3050, AUSTRALIA Viki Lapatis and /!!\ <<<<< ph: 61_3_9345 2540 fax: 61_3_9347 0852 Jennie Parker o !! \ < in:: "facs_copy@wehi.edu.au"