Re: PE-Cy5 on Facscalibur

From: Angela Schultz (Angela_Schultz@bdis.com)
Date: Thu Jun 19 1997 - 11:30:05 EST


Dear Peter Chapple,


I have asked John Wotherspoon from BD Australia to follow up with you directly
on your FACSCalibur queries. But for all other interested parties, Tom Frey, a
scientist in our Research group that worked on the FACSCalibur development adds
these comments.


"The 3-color Calibur uses a 650LP, the 4-color uses 670LP.  The maximum emission 
of a PE/Cy5 is within either of those filter ranges. PE/Texas Red is shorter and 
you may only pick up the tail (although experience suggests that works well 
enough in most cases.

BD recommends PerCP in order to avoid some of the compensation problems 
that can occur with PE/Cy5 conjugates.  They always emit some PE fluorescence 
(although often not much with a good conjugate).  The amount of PE fluorescence 
can vary from conjugate to conjugate or from vendor to vendor.  Also, it is 
fairly well known that exposure to light can bleach the Cy5 fluorochrome and 
lead to more emission in the PE channel.  Labs that are less careful about light 
exposure might see this as an increase in FL2 over the lifetime of the bottle of 
reagent."

As for the FACSCalibur manual, the Chapter on FL4 does have a diagram showing
the optical layout and the filters that are used in FL1-FL4.  If for some reason
any FACSCalibur customer did not receive the FL4 chapter when they received
their FL4 upgrade, let us know and we will get one right out to you. Also, if
you have any concerns regarding the documentation of the FACSCalibur, we would
love to hear them directly, so we can make it better to meet your needs.  Feel
free to contact me at BD directly with comments or concerns and I will forward
them on.  Thanks.

Angela K. Schultz
BDIS Marketing/FACSCalibur (Development) Program Manager
email: angela_schultz@bdis.com
_______________________________________________________________________________
Subject: PE-Cy5 on Facscalibur
From:    Peter Chapple <peterc@petermac.unimelb.edu.au> at INTERNET
Date:    6/18/97  12:33 PM


Greatings all,

I have been looking into the economics of our flow setup and discovered that I
can buy PECy5 anti CD45 for much cheaper than I currently buy PerCP anti CD45. I
proceeded to make some three color CD3/4/45 and 3/4/45 PBL's to do some
compensation with and found that I got into real trouble :(

Did something drastic go wrong with the labeling of the PBL's ??, or is there
something I need to know about PE-Cy5 on a Calibur (ie: has anyone else done
this easily/with difficulty ??)

One problem is that I have very little quality information on the Calibur - the
manual seems to make no mention of the of exactly which mirrors/filters the
Calibur uses ... does it use a 650LP filter on the FL3 detector the same as a
FacsScan does ?? (if that's the case then I'm only just picking up the *tail* of
the PE-Cy5 emission)


Peter Chapple
Melbourne AUSTRALIA


Received: from flowcyt.cyto.purdue.edu (128.210.60.31) by cc.bdis.com with SMTP
  (IMA Internet Exchange 2.1 Enterprise) id 0000998D; Wed, 18 Jun 97 12:05:10
-0700
Received: by flowcyt.cyto.purdue.edu (940816.SGI.8.6.9/930416.SGI.AUTO)
	for cyto-sendout id MAA21015; Wed, 18 Jun 1997 12:30:17 -0500
Received: from bastion.petermac.unimelb.edu.au by flowcyt.cyto.purdue.edu via
ESMTP (940816.SGI.8.6.9/930416.SGI.AUTO)
	for <cytometry@flowcyt.cyto.purdue.edu> id VAA18792; Tue, 17 Jun 1997 21:49:15
-0500
Received: (from mail@localhost) by bastion.petermac.unimelb.edu.au
(8.6.8/8.6.10) id MAA03573 for <cytometry@flowcyt.cyto.purdue.edu>; Wed, 18 Jun
1997 12:57:28 +1000
Received: from server.petermac.unimelb.edu.au(172.20.7.15) by
bastion.petermac.unimelb.edu.au via smap (V1.3)
	id sma003549; Wed Jun 18 12:56:43 1997
Received: from path5.petermac.unimelb.edu.au ([172.20.10.223]) by
server.petermac.unimelb.edu.au (8.6.8/8.6.10) with SMTP id MAA25429 for
<cytometry@flowcyt.cyto.purdue.edu>; Wed, 18 Jun 1997 12:45:58 +1000
Received: by path5.petermac.unimelb.edu.au with Microsoft Mail
	id <01BC7BE3.DF91D3C0@path5.petermac.unimelb.edu.au>; Wed, 18 Jun 1997 12:33:49
+1000
Message-ID: <01BC7BE3.DF91D3C0@path5.petermac.unimelb.edu.au>
From: Peter Chapple <peterc@petermac.unimelb.edu.au>
To: cyto-inbox
Subject: PE-Cy5 on Facscalibur
Date: Wed, 18 Jun 1997 12:33:47 +1000
MIME-Version: 1.0
Content-Type: text/plain; charset="us-ascii"
Content-Transfer-Encoding: quoted-printable



This archive was generated by hypermail 2b29 : Wed Apr 03 2002 - 11:49:52 EST