The key fluidic problem in flow cytometry is maintaining laminar flow in
order to maintain relatively uniform core diameter and position; increasing
the viscosity dramatically would probably lead to lower optimal flow
velocity, which would actually be helpful if you wanted to increase
sensitivity by increasing dwell time in the illuminating beam.  We, Los
Alamos, and others have used low flow velocities for high-sensitivity flow
cytometry (e.g., scatter measurements of viruses, fluorescence measurements
of DNA fragments).  If you used an actual flow nozzle, as with a stream in
air system, viscosity would be a problem; with a closed flow cell, it
shouldn't be. Of course, lower flow velocity will limit the number of events
you can analyze per unit time.