 In response to Fang-yao Hou, peak height measurements can under measure 
total fluorescence when the particle being measured has a cross-sectional 
diameter larger than the effective illumination spot of the laser. in 
cell-cycle analysis this situation will result in a mean channel G2/G1 ratio 
less than the theoretical 2.0 and often as low as 1.8. Therefore, pulse area 
measurements are preferred when doing DNA staining.

On Tue, 6 May 1997, Fang-Yao Hou wrote:

> 
> Dear FLOW experts:
> 
> This is to ask your opinions about using FL-height, instead of FL-area in
> cell cycle analysis. I can see typical picture of G1, S and G2/M from FL-
> height histogram as long as I acquire data in linear mode. I am wondering
> if this is an acceptable way to do cell cycle analysis without purchasing
> the pulse processing module. The cytometer I am using is the BD Vantage.
> I appreciate your responses.
> 