Re: StreamViewMonitor

From: Jim Houston (jim.houston@stjude.org)
Date: Mon Mar 10 1997 - 09:53:03 EST


On Fri, 7 Mar 1997 Pizzo@nso1.uchc.edu wrote:

> 
> 
> Colleagues,
> 
> About two years ago we acquired a used FACSTAR Plus.
> It came equipped with a stream viewing video monitor.
> I was relatively new to the field at that time and in any case
> had never worked with this particular instrument before.
> The resolution of the stream against background seemed
> poor from the start.  Recently the monitor lost all resolution
> of the stream.  I can barely make it out and its worthless from
> a practical standpoint.  I had only changed the nozzle recently
> so I tried adjusting the mirror and strobe light but to no effect.
> The machine is ten years old - is the monitor simply on its
> last legs? - the background is very greyish and the contrast
> adjust doesn't help either.  Luckily I reinstalled the microscope
> and that works OK.  Any thoughts?
> 
> Gene Pizzo/UCONN Health
> 
> 

I had attached a second monitor obtained from our biomed department to 
the existing monitor.  The monitor they gave me was no more than a 10 
inch security monitor (B/W).  These can be bought for less than $100.
It worked better than  the original.  You can do this by running a bnc 
connector through the back of the monitor and put a T on the output from 
the Camera.  IF the mointor is just junk then take the camera unit out 
and trash the monitor.

godd luck

jim