Re: FACScan cleanup - Washing solutions ???

/G=Gerhard/S=Nebe-von-Caron/OU=1890CHPE/O=TMGB.URC/@LANGATE.gb.sprint.com
Tue, 24 Oct 1995 05:14:00 -0400

We use a 10% Domestos solution made up with distilled water
and sterilefiltered. This we keep for up to 3 month in the
dark at room temperature without problems. Domestos contains
15% hypochlorit (in the UK market). It forms a precipitate
with (hard) tab water wich you would not want in the FCM.

Gerhard Nebe-v.Caron
Unilever Research, Colworth Laboratory Sharnbrook,
Bedfordshire
GB - MK44 1LQ
Tel: +44(0)1234-222066
FAX: +44(0)1234-222344
E.mail: gerhard.nebe-von-caron@urcgb.sprint.com


______________________________ Reply Separator _________________________________
Subject: FACScan cleanup - Washing solutions ???
Author: mhaury@pasteur.fr at INTERNET
Date: 23/10/95 22:34

Hello..

I'm just interested in the current washing - cleanup solution that people
use for cleaning their machines. Our FACScans systematically get clogged up
because the hypochlorite solution (supposedly 1 %) forms cristals in the
stock solution.. the only solution is to prepare fresh every day (which I'm
too lazy to do...).

Is there anyother detergent/solution that people use to cleanup that
doesn't have this problem ??

Thanks a lot,

Matthias