>I have an investigator who wants to look at intracellular distribution of a
>drug using confocal laser scanning microscopy. The drug is not fluorescent,
>so he has a monoclonal antibody (FITC conjugate) to the drug. Our question
>is, during permeablization to allow the antibody into the cells, what is the
>best way to prevent the drug from leaching out of the cells? I suppose a
>pre-permeablization paraformaldehyde fixation step would be helpful, but I
>would like to hear from anyone who has done such drug studies and knows if
>this is an effective approach.
>
Granted, I'm not exactly a confocal microscopy maven, but...

The distribution of many drugs in cells is affected by the action of pumps
and other enzymes, many of which are in turn dependent on ATP concentration
and/or on the maintenance of potential gradients across cytoplasmic and
mitochondrial membranes. Even without fixation, whatever you need to do to
make big enough holes in the cytoplasmic membrane to admit an anti-drug
antibody will transiently abolish potential gradients and allow small
molecules to leak out of the cell. Thus, what you'll see with confocal is
very likely to be artifactual, and fixation will only move things further in
that direction.

This doesn't mean that you won't get any information. For example, if you
had a drug that bound covalently or with very high affinity to DNA, you
should certainly be able to find it in the nucleus with a fluorescent
antibody, but any conclusions you might draw about cytoplasmic trafficking
of the same drug would probably be suspect.

-Howard