Re: Intracellular staining

Derek Davies (davies@icrf.icnet.uk)
Thu, 28 Dec 1995 10:13:22 +0000 (GMT)

The protocol that I find most useful for permeabilisation for the
detection of internal antigens is to use saponin. It is relatively gentle
(ie it doesn't put enormous holes in the cytoplasm or the cell membrane)
and can be used in conjunction with surface staining or DNA staining. You
may have to play about with conditions but a good starting point is to
treat cells with 0.3% saponin for about 15mins at room temperature before
doing the antigen detection and then to use 0.1% saponin in all
subsequent steps.
Be careful with the washing steps as it is easy to lose all your cells!
We have found that the permeabilisation is best at room temperature and
that the saoponin needs to be present all the time as the
permeabilisation effect can be reversed.
As with all intracellular staining washing is important, but if you are
using directly conjugated antibody, this reduces the steps and reduces
cell loss.
Hope that this is of some use.

Derek

Derek Davies
Imperial Cancer Research Fund
London
http://www.icnet.uk/axp/facs/davies/index.html

On Tue, 26 Dec 1995, Howard Ratech wrote:

>
>
> We are trying to determine if COS cells are expressing a particular
> protein after transient transfection. The protein is designed to be
> secreted but I am hoping we can find enough of it in the cytoplasm to tell
> us the experiment worked before we do a Western blot on the culture
> supernatant.We have directly PERCP conjugated mouse monoclonal antibody to
> this protein. Do you have any recommended protocol for permeablization,
> staining, washing etc.? Since these cells are quite large, is there
> anything we should be doing with the FSC and SSC, other than using linear
> amplification? We are using a BD FACSCAN with LYSIS software.
>
> Happy Holidays.
>
> Howard
>
>


Home Page Table of Contents Sponsors E-Mail Archive Web Sites

CD-ROM Vol 3 was produced by Monica M. Shively and other staff at the Purdue University Cytometry Laboratories and distributed free of charge as an educational service to the cytometry community. If you have any comments please direct them to Dr. J. Paul Robinson, Professor & Director, PUCL, Purdue University, West Lafayette, IN 47907. Phone: (765)-494-0757; FAX(765) 494-0517; Web http://www.cyto.purdue.edu , EMAIL cdrom3@flowcyt.cyto.purdue.edu


Home Page Table of Contents Sponsors E-Mail Archive Web Sites

CD-ROM Vol 3 was produced by Monica M. Shively and other staff at the Purdue University Cytometry Laboratories and distributed free of charge as an educational service to the cytometry community. If you have any comments please direct them to Dr. J. Paul Robinson, Professor & Director, PUCL, Purdue University, West Lafayette, IN 47907. Phone: (765)-494-0757; FAX(765) 494-0517; Web http://www.cyto.purdue.edu , EMAIL cdrom3@flowcyt.cyto.purdue.edu