Re: Three Colour Compensation

Marty Bigos (BIGOS@Beadle.Stanford.EDU)
29 Sep 1992 09:34:01 -0700 (PDT)

The "problem" that Alice describes is inherant in the spectrum of the dyes and
the design of the Facscan (and most other cytometers, for that matter).

FITC spectrum has significant overlap into the FL2 channel on Facscan. Moreover,
the spectrum of FITC in the FL3 channel is not the same as that of PE. So the
signal that results in the FL2 from FITC bright objects requires different
compensation than that of PE in the FL3 channel. Given the stock filters chosen
for the Facscan (585BP for FL2 and 650LP for FL3) the compensation of PE into
FL3 will be high enough to overcompensate FITC in FL3. Since there is no
compensator between FL1 and FL3 (that would go negative as well) there are no
controls that can be adjusted to correct this.

However, if the Facscan in question is not used for PI DNA analysis on FL3, a
partial solution would be to replace the FL3 filter (a 650 LP) with one better
optimized for the Cy5-PE spectrum. On our Facstar we use a 685 BP filter (which
is centered around the Cy5-PE peak). This would minimizes the FL2 to FL3
compensation thus reducing the FITC overcompensation in FL3. This filter also
allows enough PI signal to get though if it is used for live/dead cell gating;
but not enough for good quality DNA analysis.

-Marty Bigos & the gang
Stanford Shared FACS Facility


Home Page Table of Contents Sponsors E-Mail Archive Web Sites

CD-ROM Vol 3 was produced by Monica M. Shively and other staff at the Purdue University Cytometry Laboratories and distributed free of charge as an educational service to the cytometry community. If you have any comments please direct them to Dr. J. Paul Robinson, Professor & Director, PUCL, Purdue University, West Lafayette, IN 47907. Phone: (765)-494-0757; FAX(765) 494-0517; Web http://www.cyto.purdue.edu , EMAIL cdrom3@flowcyt.cyto.purdue.edu