Re: Distance between sort streams

Joseph Webster (J.Webster@centenary.usyd.edu.AU)
Fri, 13 Sep 1996 10:37:20 +1100

At 08:25 AM 12/09/96 +0300, Yash Pal Agrawal wrote:
>Hi everybody,
> I request some advice regarding sorting.
>Question: We use a FACStar for sorting. However, inspite of my best
>efforts I am not able to get the Left and Right Streams to separate much
>from the center stream. As a result the sorted cells miss the collection
>tubes. (The left and rt. stream deflection controls are turned to max).
>
>- Does the angle between the 2 charged metal plates (which is adjustable)
>have anything to do with it ?.

We had a similar problem with the FACStar Plus and found that the
deflection plates were upside down!

Yes, the angle does make some difference, but the longer part of the
plates must be UP towards the nozzle.

We run with the left & right controls at Max and the streams are about
centre of each collection tube. (cannot make it go near the outer edge
of collection tubes)

Good Luck, Joseph.

Joseph Webster (O.I.C. Flow Cytometry)
Centenary Institute of cancer Medicine & Cell Biology
Locked Bag No.6, Newtown, NSW 2042, AUSTRALIA.
Ph: 61-2-9565-6110 Fax: 61-2-9565-6101


Home Page Table of Contents Sponsors E-Mail Archive Web Sites

CD-ROM Vol 3 was produced by Monica M. Shively and other staff at the Purdue University Cytometry Laboratories and distributed free of charge as an educational service to the cytometry community. If you have any comments please direct them to Dr. J. Paul Robinson, Professor & Director, PUCL, Purdue University, West Lafayette, IN 47907. Phone: (765)-494-0757; FAX(765) 494-0517; Web http://www.cyto.purdue.edu , EMAIL cdrom3@flowcyt.cyto.purdue.edu