However, if this is still not to your liking, consider using
biotinylated anti-IFN-gamma and then coming back with streptavidin
Pe/Cy5. Biotinylated mAbs, in the context of intracellular staining have
a lot of background due to: 1.) endogenous biotin in the cell 2.) they
are just plain stickier. IFN is a very bright stainer, so even if you
don't block the endogenous biotin (see J.I., 154 4292-4301) you will
still have an acceptable level of signal.
Don't forget to eliminate sources of free biotin that will bind up your
straptavidin, such as serum.
Calman Prussin
NIAID/ NIH
>----------
>From: Ress, S, Stan, Dr
>Sent: Sunday, September 8, 1996 10:26
>To: Cytometry Mailing List
>Subject: 3-col surface marker/i.c. cytokine staing
>
>
>Hi all,
>
>We would like to stain human PBL for simultaneous dual color surface
>markers, and IFN gamma intracellular expression, using a coulter
>epics profile 2. We would like to quantitate IFN -G production by
>memory T cells and CD4/8 subsets in this way. Most of the commercial
>mca to surface markers will be FITC/PE combinations, does anyone know
>of any available IFN-gamma MCA conjugated to a third fluorochrome
>which could be used in these experiments? I understand that for I.C.
>work PE should be used, not FITC, but this won't work as PE-conjugated
>MCA will be used for surface marker as indicated.
>Any advice?
>
>Thanks,
>
>Stan.
>Stan Ress
>Clinical Immunology laboratory
>Dept of Medicine
>University of Cape Town
>Cape Town, South Africa
>
>e-mail: sress@uctgsh1.uct.ac.za
>Phone: Intern + 2721-4066201
>FAX : Intern + 2721-4486815
>
![]() |
![]() |
![]() |
![]() |
![]() |
CD-ROM Vol 3 was produced by Monica M. Shively and other staff at the
Purdue University Cytometry Laboratories and distributed free of charge
as an educational service to the cytometry community.
If you have any comments please direct them to
Dr. J. Paul Robinson, Professor & Director,
PUCL, Purdue University, West Lafayette, IN 47907.
Phone: (765)-494-0757;
FAX(765) 494-0517;
Web