Re: Linear to log channel conversion...Help

Neal Benson (nbenson@biotech.ufl.edu)
Fri, 6 Sep 96 18:25:19 edt

John,

At 08:14 PM 9/4/96 GMT, you wrote:
>Can anybody help in converting channel numbers from listmode data
>(i.e. 0 - 1023) to their correct 4 decade log value.

I believe you really want the linear representation of channel numbers when
a log amp is used to collect the data. If so, divide the channel no. by 256
and use the result as a power of 10. '256' comes from the total channels
divided by the number of decades.

i.e. c
( --- ) c is the channel number
256
10

Neal

--
Neal A. Benson
Biotechnology Program, University of Florida
E-mail: nbenson@biotech.ufl.edu   


Home Page Table of Contents Sponsors E-Mail Archive Web Sites

CD-ROM Vol 3 was produced by Monica M. Shively and other staff at the Purdue University Cytometry Laboratories and distributed free of charge as an educational service to the cytometry community. If you have any comments please direct them to Dr. J. Paul Robinson, Professor & Director, PUCL, Purdue University, West Lafayette, IN 47907. Phone: (765)-494-0757; FAX(765) 494-0517; Web http://www.cyto.purdue.edu , EMAIL cdrom3@flowcyt.cyto.purdue.edu