RE: Re: FW: BAL and clogging

vanburen%flovax.dnet.wayne.edu@rocdec.roc.wayne.edu
Tue, 9 Jul 1996 16:15:16 -0400

Simon Monard noted:
> 50u mesh sounds a bit big for a 70u nozzle.

Indeed, BDIS recommends using nozzles with a diameter that is four times
larger than the diameter of the largest particle sorted! A 200 micron tip
is available for the MacroSORT option (largest nozzle is 400 microns). For
a 70 micron tip, I use a 20 micron mesh; admittedly not quite 4:1, but at
least it gets out the visible clumps, pieces of tissue, etc. :)

/\/\/\_ Eric Van Buren, vanburen%flovax.dnet@rocdec.roc.wayne.edu
\ \ \ Karmanos Cancer Institute and Wayne State University,
\_^_/ Immunology & Microbiology. Detroit, Michigan.


Home Page Table of Contents Sponsors E-Mail Archive Web Sites

CD-ROM Vol 3 was produced by Monica M. Shively and other staff at the Purdue University Cytometry Laboratories and distributed free of charge as an educational service to the cytometry community. If you have any comments please direct them to Dr. J. Paul Robinson, Professor & Director, PUCL, Purdue University, West Lafayette, IN 47907. Phone: (765)-494-0757; FAX(765) 494-0517; Web http://www.cyto.purdue.edu , EMAIL cdrom3@flowcyt.cyto.purdue.edu


Home Page Table of Contents Sponsors E-Mail Archive Web Sites

CD-ROM Vol 3 was produced by Monica M. Shively and other staff at the Purdue University Cytometry Laboratories and distributed free of charge as an educational service to the cytometry community. If you have any comments please direct them to Dr. J. Paul Robinson, Professor & Director, PUCL, Purdue University, West Lafayette, IN 47907. Phone: (765)-494-0757; FAX(765) 494-0517; Web http://www.cyto.purdue.edu , EMAIL cdrom3@flowcyt.cyto.purdue.edu