FACScan cleaning procedures

Steven Merlin (merlin@patho.unibe.ch)
Wed, 25 Oct 1995 11:20:48 +0200

We clean the FACScan on a daily basis (or long breaks between assays) with
0.5% sodium hypochlorite (1:10 dilution of household bleach at 5.25%). This
should be prepared fresh weekly to be effective for decontamination
purposes. Run for 5-10 minutes on HIGH flow rate through sample
introduction port. This is followed with a DI water rinse (5-10min).

An excellent and inexpensive consumer household product for removing film
deposits on the interior of flow cells is a rinsing agent for automatic
dish washers. In the USA, one brand is "Jet Dry". An equivlent product in
Europe is manufactured by Lever.(Thanks to Jim Jett and John Martin of Los
Alamos for this tip). We use this for 2 minutes followed by a final 2
minute rinse of DI water.

To reduce the chance of protein from DNA and cell samples clogging tubing
or coating the flow cell, never turn the fluidics to STANDBY while the
sample is still on. If a DNA sample change will take more than 10 seconds,
placing a tube of saline or water on the uptake tube and switching to HIGH
will lessen the chance of clogs. Our system has been in operation for 2
years without any clogging by using these methods. Daily sensitivity /QC
checks have remained consistant during the same time period.

If the FACScan will not be used for more than a week, it is flushed with
sterile water and stored this way until put back into operation. Every 6
months, a 0.5% sodium hypochlorite rinse is performed as a preventive
maintenance procedure. The saline filter is removed and bypassed for this
step.

Steven Merlin
University of Bern
FACS Core Facility
Institute for Immunology and Allergy
Murtenstrasse 31
CH-3010 Bern
Switzerland
Voice: ++41-31-632-8876
FAX: ++41-31-381-8764
e-mail: merlin@patho.unibe.ch


Home Page Table of Contents Sponsors E-Mail Archive Web Sites

CD-ROM Vol 3 was produced by Monica M. Shively and other staff at the Purdue University Cytometry Laboratories and distributed free of charge as an educational service to the cytometry community. If you have any comments please direct them to Dr. J. Paul Robinson, Professor & Director, PUCL, Purdue University, West Lafayette, IN 47907. Phone: (765)-494-0757; FAX(765) 494-0517; Web http://www.cyto.purdue.edu , EMAIL cdrom3@flowcyt.cyto.purdue.edu