Re: Memory Problem with CellQuest

Geoffrey Osborne (Geoff.Osborne@anu.edu.au)
Thu, 24 Aug 1995 09:53:22 +1000

Martin,
Regarding
>I have a proplem with BD's CellQuest on a Macintosh Quadra with
>16MB RAM:
>When creating more than 6 mulitcolor Dotplots I get a message
>saying that there is not enough memory anymore.
>Is there any solution besides buying another 16MB of RAM and
>switching memory swaping to on?
>
>Any help welcomed!

This may seem like an oversimplist approach to your problem but, check that
the amount of memory you have allocated for use by CELLQuest is set to the
maximum available after that required by the system. Check the memory
currently allocated to CELLQuest using the "Get Info" option on the "File",
lock it at 10 Meg for example, and close down "unnecessary" application
which may also be running concurrently.
In our hands, on a similar system, we don't get these sort of memory errors
until we have about 32 contour plots over 4 pages, and then try and flip
things to the clipboard.

Hope this limited advise is of some use.
Geoff
======================================================================
Geoffrey Osborne | ____ __ o Ahh!
Flow Cytometry (FACS LAB) | __ `\ <,_
John Curtin School of Medical Research, | __ (*)/ (*)
Australian National University, | ==============|
CANBERRA, AUSTRALIA. | |--|
Email: Geoff.Osborne@anu.edu.au | |--|...
-----Surfing the Web?: Try http://jcsmr.anu.edu.au/facshome.html------
======================================================================


Home Page Table of Contents Sponsors E-Mail Archive Web Sites

CD-ROM Vol 3 was produced by Monica M. Shively and other staff at the Purdue University Cytometry Laboratories and distributed free of charge as an educational service to the cytometry community. If you have any comments please direct them to Dr. J. Paul Robinson, Professor & Director, PUCL, Purdue University, West Lafayette, IN 47907. Phone: (765)-494-0757; FAX(765) 494-0517; Web http://www.cyto.purdue.edu , EMAIL cdrom3@flowcyt.cyto.purdue.edu