Re: Membrane polarization

Thomas Delohery (t-delohery@ski.mskcc.org)
Tue, 30 Apr 1996 16:17:43 +0530

> We are trying to measure membrane polarization by flow cytometry
>using the probe TMA-DPH. So far we've had no luck. Is anyone else doing
>this? If so, can you tell me which probes you have found most successful
>and the specs on your polarization filters. We are using a FACS Vantage
>and "hand-me-down" polarizers (no specs). The polarizers we're using are
>yellow, and we're wondering if we were just cutting out most of the
>TMA-DPH emitted signal.

Besides your filters you also have to determine the plane of polarization
of your laser at the stream intercept. While the beam is vertically
polarized as it exits the laser, it may be in any direction after passing
through all the prisms on the Vantage. It may not even be polarized.

Fluorescence polarization measurments depend on "photoselection" - selectively
exciting molecules whose absorption dipole moments are parallel to plane of
polarization of the exciting light, ie, the laser. Molecular motion during
the excited state lifetime determines the degree of fluorescence polarization.
A freely moving fluorochrome can reorient to all possible planes during the
excited state lifetime and no fluorescence polarization is observed. An
immobilized fluorochrome at low temperature can yield highly polarized
fluorescence.

It's also extremely convenient to be able to change the lasers' polarization
for calibration purposes.

td
i've been waiting years to say that.

--
==============================================================================
 Thomas Delohery                        | Internet: t-delohery@ski.mskcc.org
 Manager, Flow Cytometry Core Facility  |    Phone: (212) 639-8729
 Memorial Sloan-Kettering Cancer Center |      Fax: (212) 794-4019
 1275 York Ave. Box 98                  |
 New York, NY    10021                  |
==============================================================================
 


Home Page Table of Contents Sponsors E-Mail Archive Web Sites

CD-ROM Vol 3 was produced by Monica M. Shively and other staff at the Purdue University Cytometry Laboratories and distributed free of charge as an educational service to the cytometry community. If you have any comments please direct them to Dr. J. Paul Robinson, Professor & Director, PUCL, Purdue University, West Lafayette, IN 47907. Phone: (765)-494-0757; FAX(765) 494-0517; Web http://www.cyto.purdue.edu , EMAIL cdrom3@flowcyt.cyto.purdue.edu