I'm wondering if your filter setup is optimal for Indo-1. It would
seem that a 505 dichroic is going to allow essentially all of the
fluorescence to enter only one of the PMTs (see Molecular Probes
catalogue, p. 114)
We do calcium measurements using Indo-1 on an ACAS 570 and the filter
setup we use is a 445 LP to split the fluorescence between the PMTs,
a 485/45 filter in front of one PMT and a 405/45 for the other PMT. THis
seems to work quite nicely.
Ray
++++++++++++++++++++++++++++++++++++++++++++++++++++++++++++++++++++++++++++++
Raymond B. Hester, Ph.D.
Flow Cytometry Lab
The University of South Alabama "Life is too short to be small."
Mobile, AL 36688-0002
Email: rhester@jaguar1.usouthal.edu Disraeli
Voice: (334) 460-6029
FAX: (334) 460-6073
++++++++++++++++++++++++++++++++++++++++++++++++++++++++++++++++++++++++++++++
On Mon, 10 Apr 1995, G Smith wrote:
> Hello,
>
> I am in the process of setting up an assay for intracellular calcium
> mobilisation using indo-1 on a FACS vantage and have run up against a
> couple of problems. If anyone could help me, it would be most appreciated.
>
> The UV excited fluorescence signal is split by a 505nm dichroic mirror
> and the two fluorescences are collected at 395+/-25nm (Fl4) and
> 525+/-25nm (Fl5). The dichroic mirror is a curious thing with a sort of
> hinge, making it infinitely adjustable. I have aligned the system with
> UV beads but get little or no signal in FL5. This may be due to the
> spectrum of the beads.
>
> With Indo-1 (3uM) loaded cells, I get a nice signal in FL4, but again no
> signal in FL5. Considering that I will be looking for an increase in FL4
> and a decrease in Fl5, this is not satisfactory. Is there an established
> setting up protocol for this assay? I get nice changes using Fluo-3
> alone, but I dont want to use this. I have a lot of noise in Fl5, and
> this may be the source of the problem. If anyone has any ideas, Id be
> pleased to hear them.
>
> Thanks
>
> Graham
> Graham
>
![]() |
![]() |
![]() |
![]() |
![]() |
CD-ROM Vol 3 was produced by Monica M. Shively and other staff at the
Purdue University Cytometry Laboratories and distributed free of charge
as an educational service to the cytometry community.
If you have any comments please direct them to
Dr. J. Paul Robinson, Professor & Director,
PUCL, Purdue University, West Lafayette, IN 47907.
Phone: (765)-494-0757;
FAX(765) 494-0517;
Web