RE: Sheath tank design to eliminate air bubbles

Eric Martz (emartz@microbio.umass.edu)
Wed, 07 Sep 1994 20:59:23 -0500 (EST)

1. Why isn't the 'final filter' taking out the bubbles?

2. The bubbles are presumably air coming out of solution due to warming
up in the afternoon. You could try degassing your sheath fluid,
e.g. putting it under partial vacuum or heating it, before use.

3. My impression is that air bubbles have a very low side scatter to
forward scatter ratio, so could be gated out on that basis. But I'd
like to hear from anyone who really knows how air bubbles look to the
cytometer.

In message Tue, 06 Sep 1994 10:56:53 -0800,
"Betsy R. Robertson" <brrob@ims.alaska.edu> writes:

>
> Late-afternoon air bubbles in sheath fluid have been a problem lately.
> They can look very much like the bacteria I am analyzing. Does anyone
> have a source or a good design for a 2-4 liter sheath tank which will
> eliminate/minimize/ reduce(?) this problem?
>
> Betsy R. Robertson
> Institute of Marine Science
> University of Alaska Fairbanks
> Fairbanks, Alaska 99775-1080
> Phone (907)474-7709
> FAX (907)474-7204 e-mail: brrob@ims.alaska.edu
>

/*- - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - -
Eric Martz, Professor of Immunology emartz@microbio.umass.edu
Dept Microbiology Voice: 413-545-2325 FAX: 413-545-1578
Morrill IVN 203, Box 35720, Univ Massachusetts, Amherst MA 01003-5720
- - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - -*/


Home Page Table of Contents Sponsors E-Mail Archive Web Sites

CD-ROM Vol 3 was produced by Monica M. Shively and other staff at the Purdue University Cytometry Laboratories and distributed free of charge as an educational service to the cytometry community. If you have any comments please direct them to Dr. J. Paul Robinson, Professor & Director, PUCL, Purdue University, West Lafayette, IN 47907. Phone: (765)-494-0757; FAX(765) 494-0517; Web http://www.cyto.purdue.edu , EMAIL cdrom3@flowcyt.cyto.purdue.edu