1. Different lots of trypan blue have different staining abilities.
The concentration may need to be adjusted for each new lot.
2. Serum or other proteins in suspension media bind trypan blue, hence
reduce the stain available to the cells. You'll need a higher concentration
of trypan if you use more protein.
/*- - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - -