Ar laser failure - summary

Karel Drbal (drbal@biomed.cas.cz)
Wed, 26 Feb 1997 07:40:17 -0800

A week ago I have asked about solving my troubles with the laser in FACSort.
Since BD technician recommended the laser replacement (in spite of the laser
current still was not so high-around 7 Amps) I asked you about your opinion.

Many of you suggested insufficient laser (cytometer) cooling:
(thermally unstable Status card; a component in the power supply is overheating
and cutting out; if a component is failing due to heat, it is likely to be some
small electronic bit rather than the laser tube).

There are no obstructions to the vent in the back of the instrument and fans
are working properly. I decided to clean the whole instrument from dust using
compressed air. The laser inside and the filter on its side was full of dust!!!
After that cleaning procedure and 1 week running the cytometer I have not
observed the failure again. Thank you,
Regards,

-- 
  _______________________________________________________________
  Karel Drbal
  Laboratory of Leukocyte Antigens
  Institute of Molecular Genetics
  Academy of Sciences of the Czech Republic
  Videnska 1083
  142 20 PRAHA 4
  Czech Republic, Europe
  voice:  +42-2-4752589
  fax:    +42-2-4727979
  e-mail: drbal@biomed.cas.cz
  WWW:    http://leuko.biomed.cas.cz


Home Page Table of Contents Sponsors E-Mail Archive Web Sites

CD-ROM Vol 3 was produced by Monica M. Shively and other staff at the Purdue University Cytometry Laboratories and distributed free of charge as an educational service to the cytometry community. If you have any comments please direct them to Dr. J. Paul Robinson, Professor & Director, PUCL, Purdue University, West Lafayette, IN 47907. Phone: (765)-494-0757; FAX(765) 494-0517; Web http://www.cyto.purdue.edu , EMAIL cdrom3@flowcyt.cyto.purdue.edu