Ray
At 3:45 pm 18/10/96, Vincent Falco wrote:
>To All:
>In the time course of performing a sort (1-2hrs.) my sorter is experiencing
>drifts in droplet breakoff(as observed by the camera screen on the
>sorter).When this is observed,the sort is stopped and the drop delay/droplet
>breakoff is checked by performing the ususal bead count matrix over several
>drop delays.Typically, the drop delay has in fact changed from its original.
>While I want to hear any and all responses as to what the problem may be(my
>performance or the sorter)assume for the time that there is absolutely no
>clogs,no air/fluidic leaks, and all electronics completely in fine tune.
>Thanks.
>Vincent Falco
>New England Medical Center
>vincent.falco@es.nemc.org
Ray Hicks
________________________________________________________________________
|University of Cambridge |Tel 01223 330149 |
|Department of Medicine |Fax 01223 336846 |
|Level 5, Addenbrookes Hospital |e-mail rh208@cus.cam.ac.uk |
|Hills Road Cambridge |Web Page/ facsmac.med.cam.ac.uk |
|CB2 |ftp server 131.111.80.78 |
|UK | |
|_________________________________|_____________________________________|
![]() |
![]() |
![]() |
![]() |