It is my experience that micro airbubbles in the sheet-fluid can cause
problems during sorting, mainly by drifting of the dropformation, thereby
reduse the purity and yield during a run. I therefore, with good success,
routinously de-gass my sheet-fluid to "no boiling" with a vacumpump and a
cold-trap between the pump and sheetfluid (not to destroy the oil-based
vacumpump). This is actually the speed-vac system adapted to fit the fluid
tank. It might be acceptable to use a standard water-jet pump to create the
vacum for de-gassing.
But, of course, if there are leakages in the system (O-rings, tubes, etc.),
this will cause troubles of unstability. These problems should be easy to
evaluate.