--------------------------------------
Date: 12/8/95 6:01 PM
To: cyto-inbox
From: Zawadzki, Jerome A.
No question about it Ned. That is going to be a very difficult combination
to work with. The efficiency of Cy3 at 488 is one problem and the PI spill
over into fl2 is the other. If your Cy3 signal is strong you may be able
compensate the PI. I would guess that the PI is for dead /live
discrimination. Have you tried the Cy3 on the instrument yet? To answer your
question about getting around this. Drop the Cy3 if you have that option. If
you have to use PI then you might try Fitc and Percp. Just some thoughts
Ned. Good Luck . Jerome
----------
From: owner-cyto-sendout
To: cyto-inbox
Subject: three color analysis
Date: Wednesday, December 06, 1995 12:05PM
I would like to do three color analysis on our FACSCAN using FITC, Cy3 and
propidium iodide. I'm told that there is too much spill over of the PI
into FL2, such that I would not get reliable results on the Cy3 staining.
Is this true. Is there anyway around this problem? Thanks. Ned Landau,
Aaron Diamond AIDS Research Center.
------------------ RFC822 Header Follows ------------------
>From flowcyt.cyto.purdue.edu!owner-cyto-sendout Fri Dec 8 17:48:47 1995 remote
from donews
Received: from crash.cts.com by donews.cts.com with smtp
(Smail3.1.29.1 #5) id m0tOEP7-00000tC; Fri, 8 Dec 95 17:48 PST
Received: from flowcyt.cyto.purdue.edu by crash.cts.com with smtp
(Smail3.1.29.1 #5) id m0tOEPy-0001CeC; Fri, 8 Dec 95 17:49 PST
Received: by flowcyt.cyto.purdue.edu (950215.SGI.8.6.10/930416.SGI.AUTO)
for cyto-sendout id RAA01831; Fri, 8 Dec 1995 17:16:18 -0500
Received: from watson.princeton.edu by flowcyt.cyto.purdue.edu via SMTP
(950215.SGI.8.6.10/930416.SGI.AUTO)
for <owner-cyto-sendout@flowcyt.cyto.purdue.edu> id LAA24085; Thu, 7 Dec 1995
11:49:32 -0500
Received: from molecular.princeton.edu by watson.princeton.edu (MX V4.1 VAX)
with SMTP; Thu, 07 Dec 1995 11:18:17 EST
Received: by molecular.princeton.edu with Microsoft Mail id
<30C73E01@molecular.princeton.edu>; Thu, 07 Dec 95 11:18:25 PST
From: "Zawadzki, Jerome A." <jzawadzki@molecular.princeton.edu>
To: cyto-inbox
Subject: RE: three color analysis
Date: Thu, 07 Dec 95 11:17:00 PST
Message-ID: <30C73E01@molecular.princeton.edu>
Encoding: 21 TEXT
X-Mailer: Microsoft Mail V3.0
X-PMFLAGS: 33554560
![]() |
![]() |
![]() |
![]() |