RE: BIG files
Doug Smoot (rin0dss@bumed30.med.navy.mil)
Wed, 5 Jan 1994 09:54:38 -0400 (EDT)
I have had a similar problem with my Elite. I discovered that the problem
in my files was in the header. When I saved my files I got an error
message saying the there was a problem saving the file for some reason I
can not remember. The FCS file format only allows for a certain size
file. I ran my files through an FCS file reading program we created and it
determined that the data area was invalid (It has been a while but I think
that is because there are not enough digits in the header space to allow
for a very large data area. What I now do is pregate the data using
the listgate feature (under the options screen on the cytometer - useful if
you are gating out non-indo loaded cells if you are running an indo
experiment). This doesn't really help with the files you have already saved
though.
In message 3 Jan 94 10:13:00 EST,
"IMVS4::KWEBER" <KWEBER%IMVS4.decnet@uv1.im.med.umich.edu> writes:
> HELP!!
>
> We have saved a number of listmode files on the Coulter Elite that are
> in excess of 10 MBytes. Unfortunately, the Elite software does not
> recognize them as valid listmode files, and the WinList software cannot
> handle displaying all 1,000,000 data points in the files. We desparately
> need to analyze these files and would appreciate it if anyone who has run
> into this problem could let us know how to work around it.
>
> Thanks!
> Kris Weber
> University of Michigan
> (313)-747-3216
> fax: (313)-764-2566
> e-mail: kweber@uv1.im.med.umich.edu
{---------------------------------------------------------------------}
Douglas Smoot
Naval Medical Research Institute, Code O6
Bethesda, MD 20889-5607
(301) 295-1121
CD ROM Vol 2 was produced by staff at the Purdue University Cytometry Laboratories
and distributed free of charge as an educational service to the cytometry community.
If you have any comments please direct them to Dr. J. Paul Robinson, Professor & Director, PUCL,
Purdue University, West Lafayette, IN 47907. Phone:(317) 494-0757; FAX (317) 494-0517; Web http://www.cyto.purdue.edu
EMAIL robinson@flowcyt.cyto.purdue.edu