RE: Distance between sort streams
James Magondu (J.MAGONDU@CGNET.COM)
Fri, 13 Sep 1996 09:36:00 -0700 (PDT)
I am sorting with the same machine on almost daily basis and the advise I
would give is to tilt the charged plates equally to make the gap between
them at the top about 4 to 5 mm. This will certainly improve the separation.
The diameter of the nozzle will also affect this separation. The bigger the
worse. I normally use a 70 um. or 80 um. The breakpoint will also matter. I
adjust the amplitude control to get a drop delay of 11.00 to 12.75 running a
drop drive frequency of 27,486 Hz. The above gives an optimum separation of
the streams.
There are certainly other factors involved and I will also keep my eyes to
my screen to hear from others.
----------
From: owner-cyto-sendout
To: Cytometry Mailing List
Subject: Distance between sort streams
Date: Thursday, September 12, 1996 8:25AM
Hi everybody,
I request some advice regarding sorting.
Question: We use a FACStar for sorting. However, inspite of my best
efforts I am not able to get the Left and Right Streams to separate much
from the center stream. As a result the sorted cells miss the collection
tubes. (The left and rt. stream deflection controls are turned to max).
- Does the angle between the 2 charged metal plates (which is adjustable)
have anything to do with it ?.
- Any other suggestions would be welcome.
Many thanks in advance.
Yash Pal Agrawal
Dept. of Clinical Chemistry,
Kuopio Univ Hospital,
Finland.
E-mail:yagrawal@messi.uku.fi
CD ROM Vol 2 was produced by staff at the Purdue University Cytometry Laboratories
and distributed free of charge as an educational service to the cytometry community.
If you have any comments please direct them to Dr. J. Paul Robinson, Professor & Director, PUCL,
Purdue University, West Lafayette, IN 47907. Phone:(317) 494-0757; FAX (317) 494-0517; Web http://www.cyto.purdue.edu
EMAIL robinson@flowcyt.cyto.purdue.edu