Re: intracellular calcium/data analysis

monard@adarc.nyu.edu
Thu, 8 Aug 1996 10:59:29 -0400

RE:
>the base line is not very stable and i am wondering is that due to the
>analysis
>or the instrument set-up. is it possible that in 10 seconds that i am missing
>the peak.
>

Frank

In cells I have looked at, the response to chemokines is pretty fast, peaks
at about 9 seconds and is pretty much back at baseline after a minute, so
you may miss the peak of the response. I also noticed that the carry over
effect from samples had a great effect, the residual chemokine in the
sample line stimulating the cells before addition of agonist. This of
course is even more pronounced with ionomycin. Flushing the sample line
with PBS for chemokines and ethanol for ionomycin for a few minutes usually
does the trick. Using an ordinary FITC filter (530/30) is better than the
485. How high are your PMT voltages?

Happy fluxing!

Simon


Home Page Table of Contents Sponsors Web Sites
CD ROM Vol 2 was produced by staff at the Purdue University Cytometry Laboratories and distributed free of charge as an educational service to the cytometry community. If you have any comments please direct them to Dr. J. Paul Robinson, Professor & Director, PUCL, Purdue University, West Lafayette, IN 47907. Phone:(317) 494-0757; FAX (317) 494-0517; Web http://www.cyto.purdue.edu EMAIL robinson@flowcyt.cyto.purdue.edu