Re: gfp ser mutant
RCherv (RCherv@mail-sh.lsumc.edu)
Tue, 07 Nov 95 22:48:18 cst
Rochelle,
We have been using this same gfp mutant with some success recently.
We started this work on a FACS Vantage with a standard FITC setup
(488nm excitation; 530/30 bp filter) and have seen good fluorescence
from transiently transfected murine cells and from cells infected with
a mutant gfp-encoding retrovirus. We are now using an improved
version of this setup, in which the 530/30 bp filter is replaced with
a 510/20 (Omega Optical). This has the dual advantage of cutting down
background from autofluorescence and increasing the signal from the
gfp. With either setup, the mutant gfp is better/brighter than the
original excited by UV in our hands. If you were able to detect
expression of the original gfp, I don't think you'll have any problems
with the mutant. If you do, let me know.
Good Luck
Rob Chervenak
CD ROM Vol 2 was produced by staff at the Purdue University Cytometry Laboratories
and distributed free of charge as an educational service to the cytometry community.
If you have any comments please direct them to Dr. J. Paul Robinson, Professor & Director, PUCL,
Purdue University, West Lafayette, IN 47907. Phone:(317) 494-0757; FAX (317) 494-0517; Web http://www.cyto.purdue.edu
EMAIL robinson@flowcyt.cyto.purdue.edu