CellQuest Crashes!

Angel_Delgado_at_USAISR1__FSHTX@ftdetrck-ccmail.army.mil
Fri, 20 Oct 95 12:20:31 EST


Hi! CellQuest Users.


About the problem with CellQuest crashing I had that problem and it
took us with, the help of BD, 4 days to figure out the exact solution.
There are actually two possible problems:

1- The SENTINEL Eve Tranceiver, which BD is having trouble
with two defective serial numbers. I only know one
serial number ending in 3024. Part# 05-10897-00

2- The Mac Aquisition Card (PLBD-BDMAC.OO2) is giving BD
troubleduring communication between the Mac and the
Cytometer. Part# 03-20061-065

NOTE: If you look though the grid on top of the Mac CPU you
will be able to see a red LED #. According to BD it
should always be 0 or 1. When mine was crashing it
showed a 7. In the system an ERROR TYPE 1 is a memory
conflicting error, in other words the board is BAD!

Angel V. Delgado
US Army Institute of Surgical Research
Biochemistry Branch
Fort Sam Houston
San Antonio, Texas 78234
(210) 221-4544


Home Page Table of Contents Sponsors Web Sites
CD ROM Vol 2 was produced by staff at the Purdue University Cytometry Laboratories and distributed free of charge as an educational service to the cytometry community. If you have any comments please direct them to Dr. J. Paul Robinson, Professor & Director, PUCL, Purdue University, West Lafayette, IN 47907. Phone:(317) 494-0757; FAX (317) 494-0517; Web http://www.cyto.purdue.edu EMAIL robinson@flowcyt.cyto.purdue.edu