> Has anyone had experience of the following software problem in BDs'
> Cellfit prog. (ver 2.0.2, Pascal 3.22): Data loads normally when the
> programme starts and can be analysed as expected, but when we
> subsequently try to load another file for analysis ('Readfile'} the
> prog. crashes irrecoverably (error code 14 if I remember rightly). I
> am using files generated both in Lysys and from a FACS440 system, but
> it doesn't seem to make any difference which I try and the fact that
> it will analyse either when the prog. first starts suggests the file
> itself is not the problem. Reloading software doesn't help. There
> is plenty of disc space.
>
> Related problem: does anyone know a way of reading single parameter
> files on Cellfit which have been generated on Lysys?
>
Richard, I believe I may have seen your problem before. It only seems to
happen when reading a Lysys file in Cellfit and relates to the file name.
Lysys allows a longer file name, so can use a longer name than Cellfit
can accept. Try re-naming your problem files with a shorter name.
Can't help on the second point, sorry.
Eric, ICRF Medical Oncology, Edinburgh UK
![]() |
![]() |
![]() |
![]() |