Re: He-Cd UV laser

Howard Shapiro (hms@shapirolab.com)
Fri, 19 Apr 1996 17:59:46 -0400 (EDT)

>we have problems with our UV laser.
>We use a Coulter Elite ESP with a He-Cd laser
>from Omnichron for DNA measurements.
>We start with a full CV measured with beads from
>Coulter - DNA check - about 2.5 to 3. After half
>a year, the CV increased to 5/6 and get worth.
>
In all likelihood, the laser has become noisy due to increased helium
pressure in the tube. This process can be slowed, but not stopped, but
running the tube for an hour or so a week, but it can't be reversed. Since
neither Coulter nor the other flow cytometer manufacturers have elected to
provide the relatively simple noise compensation circuitry which could
readily allow users to get better CV's from noisy lasers, you'll probably
have to replace the noisy plasma tube, even though power may be adequate.
However, call Omnichrome first; maybe they've found a magic fix.
--Howard


Home Page Table of Contents Sponsors Web Sites
CD ROM Vol 2 was produced by staff at the Purdue University Cytometry Laboratories and distributed free of charge as an educational service to the cytometry community. If you have any comments please direct them to Dr. J. Paul Robinson, Professor & Director, PUCL, Purdue University, West Lafayette, IN 47907. Phone:(317) 494-0757; FAX (317) 494-0517; Web http://www.cyto.purdue.edu EMAIL robinson@flowcyt.cyto.purdue.edu