Gerry-
While we are beating isotype controls to death, I should mention one
case where we find them helpful: In 3-color studies we often use one
color as a gate (e.g. CD34 plus side scatter) and the other 2 colors
to define subpopulations (e.g. HLA-DR and CD38). Since the gated population
has very different nonspecific fluorescence than the ungated cells,
and since the subpopulations are not always as clear as one would
like, a CD34- isotype-isotype control is very useful. Note that the
usual isotype-isotype-isotype control would be of no value in this
situation since it tells you nothing about nonspecificic binding
in the population of interest (in this example CD34+ cells).
@dd
Albert D. Donnenberg, Ph.D.
University of Pittsburgh
Voice: (412) 624-9596
Fax: (412) 624-9624


Home Page Table of Contents Sponsors Web Sites
CD ROM Vol 2 was produced by staff at the Purdue University Cytometry Laboratories and distributed free of charge as an educational service to the cytometry community. If you have any comments please direct them to Dr. J. Paul Robinson, Professor & Director, PUCL, Purdue University, West Lafayette, IN 47907. Phone:(317) 494-0757; FAX (317) 494-0517; Web http://www.cyto.purdue.edu EMAIL robinson@flowcyt.cyto.purdue.edu