Re:PI and 3 colour mouse lymphs

Mikael Vestberg (Mikael.Vestberg@inflam.lu.se)
Wed, 28 Feb 96 17:20:38 +0100

We have used anti CD8a-Red613 or anti-B220-Tricolour with PI. By
titrating the amount of PI used, you can separate the population from
the antibody stained ones. Usually PI staining is 5-10x brighter. In the
first case the PI+ will show up in the (red613-)corner of a PE/red613
plot, but will appear as a double positive population if you use TC as a
third colour, since less compensation is needed. We usually stain lymphs
and set amplifiers to find them around 7x10e2 on our FACSort-Cellquest
set up. PI positives are usually found above 3x10e3.

Mikael


Home Page Table of Contents Sponsors Web Sites
CD ROM Vol 2 was produced by staff at the Purdue University Cytometry Laboratories and distributed free of charge as an educational service to the cytometry community. If you have any comments please direct them to Dr. J. Paul Robinson, Professor & Director, PUCL, Purdue University, West Lafayette, IN 47907. Phone:(317) 494-0757; FAX (317) 494-0517; Web http://www.cyto.purdue.edu EMAIL robinson@flowcyt.cyto.purdue.edu