Re: Compensation versus antibody interactions

Howard Shapiro (hms@shapirolab.com)
Mon, 12 Feb 1996 20:52:37 -0500 (EST)

>>What happens to the compensation if you are running two antibodies (one
>FITC, one PE) which are interacting with each other or are recognising the
>same antigen?

Energy transfer from FITC to PE would be expected to increase the
yellow-orange and decrease the green fluorescence signal. The PE signal
would then look undercompensated, while the FITC signal would look
overcompensated. Compensation of the conventional type assumes that there
isn't significant energy transfer; how to compensate when there is is a
philosophical question.

--Howard


Home Page Table of Contents Sponsors Web Sites
CD ROM Vol 2 was produced by staff at the Purdue University Cytometry Laboratories and distributed free of charge as an educational service to the cytometry community. If you have any comments please direct them to Dr. J. Paul Robinson, Professor & Director, PUCL, Purdue University, West Lafayette, IN 47907. Phone:(317) 494-0757; FAX (317) 494-0517; Web http://www.cyto.purdue.edu EMAIL robinson@flowcyt.cyto.purdue.edu