Re: Hoechst vs FITC: possible?

Geoffrey Osborne (Geoff.Osborne@anu.edu.au)
Thu, 8 Feb 1996 10:25:08 +1100

Frank et al.
Regarding your Hoechst queries

> 3. Can the Hoechst emission spectrum be manipulated to shorter
>wavelength in any way (we have tried a longer FITC filter, 535DF15, but
>that didn't help much)? BTW, our Hoechst fluorescent beads from Flow
>Cytometry Standards give *much less* spillover!?

There is some evidence that suggests that the actual DNA/Hoechst complex
itself is very much cell type dependent, and is intimately related to the
DNA-phosphate ratio and the pH. The reference I'd point you at is Smith et
al. 1985 Flow- cytometric detection of changes in the fluorescence emission
spectrum of a vital DNA-specific dye in human tumour cells. Exptl.Cell.
res.159;37-46

With this in mind it may be worth playing with the pH, such that it is
optimized for FITC (since it's also pH dependent) and perhaps modifying the
phosphate concentration , say in the PBS in which the cells are suspended,
such that a Hoechst signal dampening occurs. I don't know if this is
possible, but it may be worth a try. I too think the spectral shift for
H333XX is concentration dependent, but as to the lowest "working" limit, I
haven't a clue.

I also recommend reading the section in James Watson's "Introduction to Flow
Cytometry" (Applications in Oncology : Therapy selection) , which could be
helpful.

Hope this helps, and if not, I've only recycled a few electrons "wastefully"!

Geoff
======================================================================
Geoffrey Osborne | ____ __ o Ahh!
Flow Cytometry (FACS LAB) | __ `\ <,_
John Curtin School of Medical Research, | __ (*)/ (*)
Australian National University, | ==============|
CANBERRA, AUSTRALIA. | |--|
Email: Geoff.Osborne@anu.edu.au | |--|...
Phone: 61 6 249 3694
FAX: 61 6 249 2595
-----Surfing the Web?: Try http://jcsmr.anu.edu.au/facshome.html------
======================================================================


Home Page Table of Contents Sponsors Web Sites
CD ROM Vol 2 was produced by staff at the Purdue University Cytometry Laboratories and distributed free of charge as an educational service to the cytometry community. If you have any comments please direct them to Dr. J. Paul Robinson, Professor & Director, PUCL, Purdue University, West Lafayette, IN 47907. Phone:(317) 494-0757; FAX (317) 494-0517; Web http://www.cyto.purdue.edu EMAIL robinson@flowcyt.cyto.purdue.edu