Re: Staining with H33342 and FITC
Hector Nolla (hector@soest.hawaii.edu)
Thu, 5 Jan 1995 08:45:08 -1000 (HST)
We routinely use hoechst(?) for staining bacterial DNA in enviromental
marine samples. Our experience has been that it is the hoechst
fluorescence which overlaps with FITC fluorescence. I think I remember
comparing the emission spectra of both and it seemed that hoechst dyes
can have a wide but dim tail of emission that overlaps with the much
brighter FITC emission. You may be able to resolve both fluorescence
better using a 2 parameter histogram that displays Log green and Log blue
fl(maybe). You could also test if FITC is excited by the UV laser by
blocking the 488nm while running your sample. I hope this helps.
Hector Nolla
Department of Oceanography
University of Hawaii
CD ROM Vol 2 was produced by staff at the Purdue University Cytometry Laboratories
and distributed free of charge as an educational service to the cytometry community.
If you have any comments please direct them to Dr. J. Paul Robinson, Professor & Director, PUCL,
Purdue University, West Lafayette, IN 47907. Phone:(317) 494-0757; FAX (317) 494-0517; Web http://www.cyto.purdue.edu
EMAIL robinson@flowcyt.cyto.purdue.edu