Re: Sheath tank design to eliminate air bubbles

Dean Hewish (deanh@tigger.mel.dbe.CSIRO.au)
Thu, 8 Sep 1994 13:17:50 +1000

For critical applications like sorting, Coulter recommend de-gassing the
sheath fluid by connecting the sheath tank pressure line to the vacuum
connection for the waste tank for 30 minutes or more, with occasional
agitation to dislodge bubbles, before using the sheath. This may not be
possible on all instruments, but vacuum treating the fluid before putting
it in the sheath tank might eliminate the problem.

Dean R. Hewish, Cell Biologist & Flow Cytometrist.
CSIRO Biomolecular Engineering, 343 Royal Parade, Parkville, 3052,
Victoria Australia.


Home Page Table of Contents Sponsors Web Sites
CD ROM Vol 2 was produced by staff at the Purdue University Cytometry Laboratories and distributed free of charge as an educational service to the cytometry community. If you have any comments please direct them to Dr. J. Paul Robinson, Professor & Director, PUCL, Purdue University, West Lafayette, IN 47907. Phone:(317) 494-0757; FAX (317) 494-0517; Web http://www.cyto.purdue.edu EMAIL robinson@flowcyt.cyto.purdue.edu