A) Bromodeoxyuridine/Propidium Iodide
The classical method for the analysis of cell cycle distribution is
the flow cytometric measurement of DNA content which can simultaneously
determine the incorporation of Bromodeoxyuridine (BrdU). The procedure
requires that DNA is partially denatured to expose incorporated BrdU to
a specific antibody. Denaturation is necessary because antibodies developed
so far bind only to BrdU in single-strand DNA. The remaining undenatured
DNA is then stained with Propidium Iodide (PI). Green fluorescence from
the fluorescein-conjugated antibody is a measure of BrdU incorporation.
Red fluorescence from the PI is a measure of DNA. The protocol described
here uses high-molarity HCl for the denaturation of DNA. Furthermore, this
method may be utilized either for unfixed or for fixed cells in suspension.
B) Cyclins/Propidium Iodide
Cyclins are key components of the cell cycle progression machinery.
In particular, the expression of cyclins D, E, A and B1 provides new cell
cycle landmarks that can be used to subdivide cell cycle into several distinct
subcompartments. In this procedure cyclins expression is detectable using
specific monoclonal antibodies (mAbs), and is analysed in respect to DNA
content.
Generally, the peak of expression of cyclin D1 can be detected in early G1, the peak of cyclin E is typical of G1/S transition, the peak of cyclin A can be detected during G2/M phases and cyclin B1 is typical of late G2/M. Using this method, compared to the above mentioned protocol, it is possible to distinguish G0 from G1 and G2 from M phases. However, it is necessary to keep in mind that not all cell types behave in the same manner (for example, cyclin D1 is detectable not only in G0/G1 but also in G2/M, even if in a very few cell types).
C) TUNEL/Propidium Iodide
One of the most used protocol for the determination of apoptosis in
the different phases of cell cycle is the enzymatic in situ labeling
of apoptosis-induced DNA strand breaks (TUNEL). Terminal deoxynucleotidyl
transferase (TdT) have been used for the incorporation of fluorescein-labeled
nucleotides to DNA strands breaks in situ. DNA content is revealed
by red fluorescence from PI. In order to have more details, see the Chapters
related to TUNEL technique.
D) F-Actin/Propidium Iodide
The analysis of apoptotic cells and estimation of their cell cycle
specificity is also possible using a recent method. This is based on identification
of apoptotic cells which have modified their cytoskleton and their DNA
content. In specific, paraformaldehyde (PFA) fixation followed by staining
of F-actin with fluorescein-conjugated phalloidin and of DNA with PI, are
used. Furthermore, this procedure may be utilized also for adherent cells.
2. PROTOCOLS
1. Cells (1x106/mL) are incubated with BrdU 10 mM
at final concentration, for 30 min at 37 °C in controlled atmosphere.
2.Wash twice at 500 g for 1 min using the washing buffer.
3. Resuspend in 0.5 mL of washing buffer and 0.5 mL of HCl 4 M.
4. Mix accurately and incubate for 30 min at room temperature.
5. Wash once as in step 2.
6. Resuspend in 1 mL of Borax buffer.
7. As in step 5.
8. Resuspend in 200 mL of washing buffer
and label with 5 mL of mAb ant-BrdU.
9. Incubate for 1 hour at 4 °C in the dark.
10. As in step 5.
11. Resuspend in 200 mL of washing buffer
and label with 4 mL of goat-anti-mouse FITC-conjugated
antibody.
12. Incubate for 30 min at 4 °C in the dark.
13. As in step 5.
14. Resuspend in 200 mL of washing buffer
and 200 mL of PI buffer.
15. Incubate for 15-30 min at 4 °C in the dark.
16. Analyse with flow cytometer equipped with a 488 nm argon laser.
A.3. COMMENTARY
A.3.1 Background information
In this procedure fixed cells by 4% PFA in Phosphate Buffer Saline
(PBS) can be utilized. In this case to wash cells once in PBS before to
start at step 1 is necessary.
Moreover, both direct and indirect immunofluorescence can be used. The BrdU incorporation is more evident using the indirect method.
B) Cyclins/PI protocol
B.2.1 Materials
PFA (A2), Triton X-100 (A2), PBS, mouse serum (A2), mAbs anti-cyclins
(A2), goat-anti-mouse-FITC (A2), PI (A2), GM+EDTA buffer (A1).
In the case of E, A and B1 cyclins, cells (2x106/mL) are fixed by 70% ethanol:
a) put all the reagents in ice;
b) count and centrifuge cells at 375 g for 5 min;
c) resuspend accurately the pellet in 1mL GM+EDTA buffer ;
d) add gently 3 mL of 96% ethanol vortexing samples and working in
ice;
e) store the fixed samples at 4°C.
In the case of D cyclins, cells are fixed by 1% methanol-free formaldehyde:
a) put all the reagents in ice;
b) count and centrifuge cells at 375 g for 5 min.;
c) resuspend gently the pellet in 1 mL of 1% formaldheyde in PBS for
15 min in ice;
d) wash in ice-cold PBS as in step b;
e) add 1 mL of 70% ethanol;
f) store the fixed samples at 4°C.
1. Wash with ice-cold PBS.
2. Resuspend the pellet in 1 mL of 0.25% Triton X-100 in PBS, vortexing samples and working in ice (for 5 min in the case of D, A and B1 cyclins, for 10-20 min in the case of E cyclin).
3. Wash in ice-cold PBS as in step b.
4. Incubate cells with 150 mL of mAb anti-cyclin (diluited at the concentration of 2.5 mg/mL in PBS + 1% of normal goat serum) overnight at 4°C.
5. As in step 3.
6. Incubate for 1 hour with 150 mL secondary mAb goat-anti-mouse-FITC (1.3 mg/mL) diluited 1:50 in PBS + 1% normal goat serum.
7. As in step 3.
8. Incubate for 4 hours at room temperature and in the dark with 1 mL of 2.5 mg/mL PI in PBS (+ 12.5 mL of 1 mg/mL RNAsi) or incubate overnight at 4°C with 0.8 mg/mL PI in PBS (+ 12.5 mL of 1 mg/mL RNAsi).
9. Analyse with flow cytometer equipped with a 488 nm argon laser.
B.3. COMMENTARY
B.3.1 Background information
The critical steps in the methodology are cell fixation, permeabilization
and the concentrations of anti-cyclin mAbs. For most cyclins optimal fixation
is 70% ethanol. This treatment preserves cyclins, lowering the background,
non-specific cell fluorescence and resulting in an improved signal-to-noise
ratio of the cyclin specific fluorescence. Detection of D cyclin, however,
requires fixation in formaldehyde. As far as anti-cyclin mAb concentration
is concerned, 2.5 mg/mL is optimal for most
cells. Anyway, to test the best concentration for each experimental model,
is recommended.
B.3.2 Anticipated results
In this procedure, a negative control sample, which contains only the
secondary FITC-mAb, is necessary.
B.3.4 Key references
2. Faretta, M., Bergamaschi, D., Ronzoni S., D’Incalci, M., Erba, E.
1997. Diferences in cyclin B1 expression in cell cycle blocked in the G2/M
phase after treatment with anti-cancer agent. A new three parametric flow
cytometry analysis. Proceedings of the XIV National Italian Meeting of
Cytometry.
3. Gong, J., Traganos, F., Darzynkiewicz, Z. 1993. Simultaneous analysis of cell cycle kinetics at two different DNA ploidy levels based on DNA content and cyclin b measurements. Cancer Res. 53: 5096.
4. Gong, J., Li, X., Traganos, F., Darzynkiewicz, Z. 1994. Expression
of G1 and G2 cyclins measured in individual cells by multiparameter flow
cytometry: a new tool in the analysis of the cell cycle. Cell Prolif.
27: 357.
5. Gong, J., Traganos, F., Darzynkiewicz, Z. 1995. Discrimination of G2 and mitotic cells by flow cytometry based on different expression of cyclins A and B1. Exp. Cell Res. 220: 226.
6. Widrow, R.J., Rabinovitch, P.S., Cho, K., Laird, C.H. 1997. Separation
of cells at different times within G2 and mitosis by cyclin B1 flow cytometry.
Cytometry 27: 250.
C) TUNEL/PI protocol
C.2.1 Materials
formaldehyde (A2), ethanol, reaction mixture (A1), TdT buffer (A1),
Bio-16-dUTP (A2), TdT enzyme (A2), staining buffer (A1), SSC buffer (A1),
BLOTTO (A2), Avidin-FITC (A2), Triton X-100 (A2), PI (A2), DNAase buffer
(A1).
C.3. COMMENTARY
C.3.1 Background information
This procedure is complex and not always good results are obtained.
Thus, the use of commercial kits such as ApoTagTM (Oncor, Gaithersburg,
MD, USA) and "In situ cell death detection Kit" (Boeringer-Mannheim, Germany),
is highly recommended.
2. Gorczyca, W.,Tuziak,T., Kram, A., Melamed, M.R., Darzynkiewicz, Z.
1994. Detection of apoptosis-associated DNA strand breaks in fine-needle
aspiration biopsies by in situ end labeling of fragmented DNA. Cytometry
15: 169.
3. Li, X., Darzynkiewicz, Z. 1995. Labelling DNA strand breaks with BrdUTP. Detection of apoptosis and cell proliferation. Cell Prolif. 28: 571.
D) F-Actin/PI protocol
D.2.1 Materials
PFA (A2), PBS, Triton X-100 (A2), sodium borohydride, FITC-phalloidin
(A2), PI (A2).
1. Cells are fixed in 1 mL of 1% PFA for 30 min on ice.
2. Wash with 0.1% Triton X-100 in PBS, and incubate with 0.1% sodium
borohydride in PBS (pH 8.0) for 30 min.
3. Wash at 200 g for 5 min.
4. Incubate with 20 mL of FITC-phalloidin
(0.01-10.0 mg/mL) for 1 hour at room temperature
(or overnight at 4°C).
5. As in step 3.
6. Resuspended in 1 mL of a 5-50 mg/mL PI
in PBS and incubate for 30 min at 37°C.
7. Analysed with flow cytometer equipped with a 488 nm argon laser.
D.3. COMMENTARY
D.3.1 Background information
Using this protocol, the acquisition and analysis of the samples is
particularly important. Apoptotic and non apoptotic cells are distinguished
on the basis of the green fluorescence and the side scatter. Apoptotic
cells have high side scatter and low FL-1 (1). The analysis of DNA content
is relative to the different regions of apoptotic and non apoptotic cells.
ii) wash at 200 g for 5 min (continue to step 1).
D.3.3 Time considerations
The protocol is relatively simple and fast, in particular 2 hours and
half are basically necessary.
Appendix 1: Stock solutions
| Solution | Preparation | Storage |
| A.Washing buffer | 0.5%Tween 20 in PBS | 4°C |
| A.Borax buffer | 0.1M Borax (Sodium tetraborate-10-hydrate) | RT |
| A. PI buffer | 3.4mM Trisodium Citrate, 9.65mM NaCl, PI 20 mg/ml, 0.03% Nonidet P-40 in H2O | 4°C |
| B.GM+EDTA buffer | glucose 1.1 g/L, NaCl 8 g/L, KCl 0.4 g/L, Na2HPO4.2H2O 0.2 g/L, KH2PO4 0.15 g/L, EDTA 0.2 g/L | 4°C |
| C.reaction mixture | 50 mL of solution was composed by: 37.8 mL of deionized water + 5 mL of TdT buffer (10X), + 5 mL of CoCl2 (25mM), + 2 mL of Bio-16-dUTP + 0.2 mL TdT enzyme | 0°C |
| C.TdT buffer (10X) | 1M Na cacodylate (pH 7.0), 1mM dithiothreitol, 0.5 mg/mL serum albumin | 4°C |
| C.staining buffer | 100 mL of solution was composed by: 54.2 mL of deionized water + SSC buffer (20X), + 20 mL of BLOTTO (25%) + 0.7 mL Avidin-FITC (160X), + 0.1 mL of Triton X-100 | 4°C |
| C.Avidin-FITC 160X | 1 mg Avidin-FITC in 250 mL PBS. Then diluit 1/10 in deionized water to have 160X stock | 4°C |
| C.SSC buffer (20X) | 0.3% sodium citrate, 3M NaCl (pH7.0) | RT |
| C.DNAase buffer | 20 ng/mL DNAasi, 10mM TRIS-HCl (pH 7.4), 10mM NaCl, 5mM MgCl2, 0.1mM CaCl2, 25mM KCl | 0°C |
Appendix 2: Reagents
| Avidin-FITC
Sigma Aldrich |
A2910 |
| anti-BrdU antibody
Becton Dickinson |
347580 |
| Bio-16-dUTP (50nmol/50mL)
Boehringer Mannheim |
1093070 |
| BLOTTO (dry non-fat milk)
Bio-Rad |
170-6404 |
| Borax
Riedel-Dehaen |
31457 |
| BrdU
Sigma Aldrich |
B5002 |
| DNAasi
Boehringer Mannheim |
776785 |
| FITC-phalloidin
Sigma Aldrich |
P5282 |
| formaldehyde
BDH |
10113 |
| goat-anti-mouse-FITC antibody
Becton Dickinson |
349031 |
| Anti A cyclin antibody
Santa Cruz Biotechnology |
sc-239, sc-596, sc-751 |
| Anti B1 cyclin antibody
Santa Cruz Biotechnology |
sc-245, sc-752, sc-595,
sc-594 |
| Anti D1 cyclin antibody
Santa Cruz Biotechnology |
sc-6281, sc-246, sc-450,
sc-717, sc-753, sc-618 |
| Anti E cyclin antibody
Santa Cruz Biotechnology |
sc-247, sc-198, sc-481 |
| mouse serum
Caltag |
10410 |
| Nonidet P-40
Sigma Aldrich |
N6507 |
| paraformaldehyde
Sigma Aldrich |
P6148 |
| PI
Sigma Aldrich |
P4170 |
| Bovine serum albumin
Sigma Aldrich |
B7276 |
| Triton X-100
Sigma Aldrich |
T9284 |
| TdT enzyme (25 U/mL)
Boehringer Mannheim |
220582 |
| Tween 20
Merk-Schuchardt |
822184 |
Appendix 3: Equipment
| Flow Cabinet TC60 | Gelaire |
| Flow Cytometer FACScan | Becton Dickinson |
| Incubator CO2-AUTO-ZERO | Heraeus |
| Minifuge RF | Heraeus |
| Pipetman P20, P200, P1000 | Gilson |
| Vortex Vibrofix VF1 Electronic | Janke & Kunkel-Ika
Labortechnik |
| Water Bath D8 | Haake |