Where is the inaccuracy?
•Consider the 14 bit data (16,384 channels)
•The smallest signal on a 0-10volt scale will be 610 uV per channel
•Thus a 1 channel change produces a value of 1220 uV or  100% possible error at the low end – since the bottom 10mV of this scale is represented by channels 1-16, the voltage at channel 16 is 9765 mV or at ch# 15 is 9765 uV or an error of about 6%
•This is an unacceptable high error at the low end so we must try to digitize at a higher bit rate say for example 16 bits (65, 536)
•Now the same range as above a 1mV signal will appear in ch# 7 and a 10 mV signal in Ch# 65 giving an error of 6% at the bottom end and only 2% at the top end
•