Fixation issues Submitted by

"Kennedy, E. (Elaine)" E.Kennedy@organon.co.uk on 12 February 2003 22:52

 

Hi Elaine

 

I will reply directly to you instead of the list because this may

sound like a commercial endorsement but it is not - Caltag's Fix and

Perm is the best.  I have used it on human and mouse PB and BM and on

many cell lines with many monoclonal antibodies and it always works,

maintains the scatter patterns well, and does not result in much cell

loss.  It is a 2-step procedure with a 15 minute incubation and a

wash step between the 2 reagents.

 

As in any new procedure, you will have to try it out with your application.

--

Barbara J Taylor

----------------------------------------------------------------------------------------

 

Elaine,

 

In my experience it is true that you may have to try several methods until you find the optimal technique.  However, in human cells we have found that Caltag's Fix & Perm Kit works almost every time.  Rarely have I had a problem using this.

(Oh - and it's really easy!  Way easier than ethanol fixation!)

 

Good Luck!

Abby Kelliher

 

-----------------------------------------------------------------------------------------------

 

Hi Elaine!

 

The most simple and in most cases perfect fixation procedure for

intracellular proteins is using cold 70% Ethanol. It works for most

proteins, you can store your samples in the fridge for a year or longer

without any changes in antigen distribution, you can do a simultaneous DNA analysis for cell cycle etc. Due to the possibility of long time storage you can analyse all the samples from an experiment at the same time which improves comparability. If your protein is very small there may be a problem with loosing it, although most, even small proteins are precipitated. There have been some cases described in the literature that the conformation of some antigens may change so that the antibodies will not recognize them anymore, but these cases are very rare. The only thing you have to take care of is the possible formation of aggregates during the fixation procedure. Here is the protocol I use for intracellular immunofluorescence:

 

Centrifuge your cells in a tube with a conical bottom, discard the

supernatant,  resuspend the cells completely in the remaining drop. Now

comes the tricky part: Put the tube on a vortex machine, keep it shaking while adding the ethanol dropwise and slowly. This is to avoid aggregate formation. Use ~1ml ethanol for 5x10e6 cells. Keep in the fridge until use.

 

Before staining centrifuge again, discard supernatant, resuspend cells in remaining droplet. Add 1ml Buffer (0.1%Tris.HCl, 0.1% Triton X100, 2mM MgCl2 pH7.4) the same way as when you added the ethanol (dropwise and slowly while vortexing). Wash again with the same buffer, again taking care to avoid aggregates. Wash again with trisbuffer including 20% FCS or 2% BSA and incubate for 30 min to block unspecific binding sites. Wash again blocking buffer and include your antibodies. After the last staining step was with buffer w/o FCS or BSA and analyse.

 

 

Hope this helps.

 

With kind regards,

Nicole

--------------------------------------------------------------------------------------------------

 

Hi Elaine,

 

I can recommend CALTAG Fix and Perm for intracellular staining and I

have used this product in both diagnostic and research laboratories.

The reagents are easy to use and also allow concurrent surface

staining. The FSC/SSC presentation does not alter significantly

compared to other permeabilising reagents.

 

I have no commercial interest in CALTAG products and I offer this

information as my opinion only.

 

Good luck!

Cathy

 

------------------------------------------------------------------------------------------------

 

Elaine,

 

In my experience you may need to look at different fixatives as well as

permeabilization agents.

 

We are looking at human white blood cells and we need both good

intracellular perm as well as good light scatter properties (for differentiating the WBC subsets). This may not be a concern for you since you're using a cell line.

 

If you're looking at something novel intracellularly, use something known first to validate your fix/perm . For example, with our

WBC work we look for good myeloperoxidase labeling in granulocytes to

validate the fix/perm.

 

I hope this helps.

 

Peter

Peter Lopez

 

-----------------------------------------------------------------------------------------------------------------------

Hi Elaine,

 

Ethanol is a penetrating fixative and conditions of use  including cell type will govern the distance of effect into a cell, for most cells it is a total fixation.

 

For DNA staining with dyes this is fine but for Ab you have to know they are against the fixed epitope.

 

For most Ab  used in Flow Cytometry it is preferable to Fix the cell membrane with paraformaldehyde then use a detergent to permiabilise said memebrane.

 

FIX

 

Paraformaldehyde fixative solution PFM

         4 % (w/v) paraformaldehyde

         3 % (w/v) sucrose  ( I never used sucrose in my fixation mix

only in the Hypertonic solution desined for membrane

shedding to recover nucli. You could try it for comparison)

         in PBS pH 7.4.

         Store at 4°C in the dark for a few days.

 

PERM

 

Triton permeating solution

         1 % (v/v) Triton-X100

         in PBS pH 7.4.

         Store at room temperature for a few weeks.

 

Saponine permeating solution 

   0.1   % w/v

         in PBS pH 7.4.

         Store at 4°C for a few days.

 

 

Nonident permeating solution

         0.5 % (v/v) Nonidet P-40

         in PBS pH 7.4.

         Store at 4°C for few days.

 

N-OctylGlucosamine  (NOG) permeating solution

         0.74 mg/l  (w/v) NOG  (This is the Critical Micelle

Concentration for this detergent)

         in PBS pH 7.4.

         Store at 4°C for few days.

 

The problems with " home brew" are reproduction, QC and shelf life so I went to a commercial product.  We sell the  IntraPrep kit for this which has a PFM fixation and Saponin Perm.  50 test IM2388 and 150 test IM2389 As a positive control to demonstrate perm and staining we have anti-Tubulin conjugated with FITC part 6607113

 

Regards

 

Martin

 

-----------------------------------------------------------------------------------------------

 

Hi Elane,

 

I did some intracellular staining in thymocytes some time ago. I fix my

cells with formaldehy solution.

 

My experience was that is was critical, how fresh the formaldehyd solution was. I took 2% formaldehy and end up with 1% end concentration for fixation.

 

I attach a protocol in pdf format and hope this helps a bit

 

Good luck

 

Steffen

 

 ====================

Dr. Steffen Schmitt

----------------------------------------------------------------------------------------------------------

 

Elaine,

 

I have used the Caltag fix and Perm kit for staining of intercellular

parasites with SOME luck, but I had better luck with 70% MeOH (Never

tried EtOH).  I would appreciate hearing relpies you get to this because I will likely try again soon.

 

Mike

 

-----------------------------------------------------------------------------------------------------------

 

Elaine,

 

We've actually pondered this same question at Cell Signaling since our

customers use a variety of different fix&perm methods.  We have settled on a protocol that involves fixation in 1% formaldehyde for 10 min at 37C, then permeabilization in 90% methanol.  This protocol works very well for every antibody we tried.  Some of our collaborators recently compared this protocol to a saponin-based fix&perm kit from Invitrogen and screened with a number of antibodies.  All of the antibodies worked well with the aldehyde/methanol, but over half of them did not work at all with the saponin-based kit.  The only down-side to methanol permeabilization is that the scatter characteristics of the cells are not as clear, so it may be difficult to pick out different cell types in a heterogeneous suspension on a scatter plot.  This isn't an issue for researchers like you that are working with cell lines.

 

For a detailed copy of our protocol, please visit our website

(http://www.cellsignal.com), click on Support, then Research Protocols, and finally on Flow.  Please feel free to contact me if you have any questions.

 

Best of luck,

--Randy Wetzel

 

 

---------------------------------------------------------------------------------------------------------------------------