SNARF issues

 

Karim Y. Vermaelen, MD, PhD

Cellular and Molecular Biology Research Laboratory

Building 37 - Room 1096

NASA - Lyndon B. Johnson Space Center

2101 NASA Parkway

Houston, Texas 77058

 

tel (+281 24)4-1993

fax (+281 48)3-3058

 

Original Question

 Has anyone here some experience in loading cells with SNARF-1? We need to label human dendritic cells with this dye. I have a protocol that works very well for CFSE / CFDAse. As far as I understand, the labeling mechanism should be similar, but I don't know whether the same concentration, incubation times would apply to the SNARF-ester.Any suggestions would be greatly appreciated!

------------------------------------------------------------------------------------------------------

Thanks to everyone who answered my question on SNARF-1 labeling. A  compilation of answers is provided below. I ended up using exactly the  same protocol as for CFSE ( 20 x 10E6 DCs/ml, 5 microM SNARF-1 final  conc., incubate 5 min at RT in HBSS, wash 3x with serum-rich medium)  -it worked perfectly (intensity, viability).

 

Responses:

 

Dear Karim - I have used Snarf before ... although it has been a long time so my memory might be wrong.  However, i'm fairly certain i found that the viability with Snarf was similar to the CFSE concentration we were  using (we got these both from Molecular Probes).... so i label with the CFSE/or/SNARF in PBS... at a range from aroudn 0.5uM to 2uM..   although in my  experience, going higher causes cells (T cells) to start dying...

 

hope that might help-

Mark

 

Hi Karim,

 

We have previously used SNARF-1 in parallel with CFSE to stain PBMC and the staining methods were the same (10 mins at 37'C, quench staining with serum followed by washes). You might want to test the toxicity of SNARF by reversing the dyes on your stained populations to compare cell viability. CFSE is safe up to 5uM and we've used SNARF at the same concentration  and it seems okay. The intensity isn't quite as intense but I wouldn't  recommend anything above 5uM.

 

Good luck and cheers.

--

Socheata Chea

Department of Microbiology & Immunology

The University of Melbourne

Parkville, VIC 3010, Australia

Phone: 61 3 8344 9938

Fax: 61 3 8344 3846

 

 

Hi Karim,

 

First thing is to order the right product - SNARF-1-AM, preferably  'special packaging' (Molecular Probes C-1272).  I follow largely the method of Chow and Hedley in 'Current Protocols in Cytometry' (1997).  Another good reference is Wieder et al Cytometry 14(8), 916-21, 1993.  Old, but good for technique.  Like them, I find time and temp of labelling has little effect and use room temp for 30 min (RT to avoid shift of pH with temp change on the FACS).  Buffers I would suggest to try are medium (no serum), PBS or HeBS (145mM NaCl, 5mM KCl, 1mM CaCl2, 0.5mM MgSO4, 5mM glucose, 10mM Hepes pH 7.4).  I label with 5ug/ml of SNARF, the concentration does seem to matter, if you don't have a red pellet they aren't labelled

   

To read, we excite at 514nm to avoid GFP interference but 488 is fine, and measure 640nm/580nm emission.  You need to do a calibration curve with each expt.  I calibrate using the nigericin method of Chow and Hedley.

 

Hope this answers your question, best of luck.

 

Kellie Tainton

kellie.tainton@petermac.org

 

 

Hi Karim,

 

when you’re using SNARF®-1 carboxylic acid, acetate, succinimidyl ester, the same conditions as for CFDA-SE, i.e. 1.10 ug/ml in PBS and approx. 30 minutes, should give good results.

 

   Hope this helps,

 

Arne

 

Hi, Karim:

We have somebody in the lab who had success with labelling myoblast cells and bone marrow SP cells with 20uM CFSE(from Molecular Probe) (in Opti-MEM I medium without serum) at 37C for an hour. I was able to detect the SNARF staining later using FACS.

Nan

 

Nan Jiang

 

Department of Internal Medicine

Cardiology Division

UT Southwestern Medical center at Dallas

6000 Harry Hines Blvd.

Dallas, TX  75390-8573

 

 